PM281X: Switching between 110V and 220V: Difference between revisions
(Remove stub)
Line 5: Line 5:
|Difficulty=1. Easy
|Difficulty=1. Easy
== Problem description ==
Revision as of 04:26, 23 February 2026
Device PM281X
Affects part(s)
Needs equipment Screwdriver
Difficulty ◉◌◌◌ Easy
Type Teardown
If you have bought a Philips/Fluke PM2811, PM2812, or PM2813 power supply from a country who's electric grid runs at a different voltage, you may want to modify your power supply, to match your electrical grid (instead of using a transformer).
Black wire, 110V position. (Figure 1)
Black wire, 220V position. (Figure 2)
Blue wire, 110V position. (Figure 3)
Blue wire, 220V position. (Figure 4)
These power supplies are internally configured to one of the following ranges:
  • 94V .. 130V (a.k.a. nominally 110V)
  • 190V .. 255V (a.k.a. nominally 220V)
To switch the voltage of the power supply between 110V/220V configurations you have to take the following actions:
  • Move the black jumper wire on the "Input Module" PCB, to the correct jumper. (this is marked via silkscreen on the PCB)
    • (Note: the service manual forgets to mention this)
    • See Figure 1 and Figure 2
  • The blue (bl) wire of the cable loom DB1 must be connected to
    • X4106 for 110V operation
    • X4102 for 220V operation
    • You can see the location of these pins on page 63 of the service manual [1]
    • Also see Figure 3 and Figure 4 (note that the PCB from the factory has no text next to these pins, you can only identify them based on relative position)
  • The fuses must be replaced. The required fuse current is different for each voltage configuration, and also depends on the number of modules the power supply is configured with
    • Note that number of modules is not the same as number of outputs. Some PM2812 have 2 power modules, while some have 3.
  • Please write on the back of the unit the current supported voltage.
Fuses
Required fuse value
Fuses 2x: 1 Module
PM2811
2 Modules
PM2812/[0,1,2]
3 Modules
PM2812/[3,4]; PM2813
110V 2.5AT 4AT 6.3AT
220V 1.25AT 2.5AT 3AT
