| (5 intermediate revisions by the same user not shown) | |||
| Line 2: | Line 2: | ||
|Device=PM281X | |Device=PM281X | ||
|Needs equipment=Screwdriver | |Needs equipment=Screwdriver | ||
|Type= | |Type=General | ||
If you have bought a Philips/Fluke PM2811, PM2812, or PM2813 power supply from a country who's electric grid runs at a different voltage, you may want to modify your power supply, to match your electrical grid (instead of using a transformer). | If you have bought a Philips/Fluke PM2811, PM2812, or PM2813 power supply from a country who's electric grid runs at a different voltage, you may want to modify your power supply, to match your electrical grid (instead of using a transformer). | ||
[[File:Pm281x_black_wire_110v_position.jpg|thumb|Black wire, 110V position. (Figure 1)]] | |||
[[File:Pm281x_black_wire_220v_position.jpg|thumb|Black wire, 220V position. (Figure 2)]] | |||
[[File:Pm281x_blue_wire_110v_position.jpg|thumb|Blue wire, 110V position. (Figure 3)]] | |||
[[File:Pm281x_blue_wire_220v_position.jpg|thumb|Blue wire, 220V position. (Figure 4)]] | |||
These power supplies are internally configured to one of the following ranges: | These power supplies are internally configured to one of the following ranges: | ||
* 94V .. 130V (a.k.a. nominally 110V) | * 94V .. 130V (a.k.a. nominally 110V) | ||
* 190V .. 255V (a.k.a. nominally 220V)To switch the voltage of the power supply between 110V/220V configurations you have to take the following actions: | * 190V .. 255V (a.k.a. nominally 220V) | ||
To switch the voltage of the power supply between 110V/220V configurations you have to take the following actions: | |||
* Move the black jumper wire on the "Input Module" PCB, to the correct jumper. (this is marked via silkscreen on the PCB) | * Move the black jumper wire on the "Input Module" PCB, to the correct jumper. (this is marked via silkscreen on the PCB) | ||
** (Note: the service manual forgets to mention this) | ** (Note: the service manual forgets to mention this) | ||
** See Figure 1 and Figure 2 | |||
* The blue (bl) wire of the cable loom DB1 must be connected to | * The blue (bl) wire of the cable loom DB1 must be connected to | ||
** X4106 for 110V operation | ** X4106 for 110V operation | ||
** X4102 for 220V operation | ** X4102 for 220V operation | ||
** You can see the location of these pins on page 63 of the service manual <ref>https://www.eevblog.com/files/philips_pm2811_pm2812_pm2813_sm.pdf</ref> | ** You can see the location of these pins on page 63 of the service manual <ref>https://www.eevblog.com/files/philips_pm2811_pm2812_pm2813_sm.pdf</ref> | ||
** Also see Figure 3 and Figure 4 (note that the PCB from the factory has no text next to these pins, you can only identify them based on relative position) | |||
* The fuses must be replaced. The required fuse current is different for each voltage configuration, and also depends on the number of modules the power supply is configured with | * The fuses must be replaced. The required fuse current is different for each voltage configuration, and also depends on the number of modules the power supply is configured with | ||
** Note that number of modules is not the same as number of outputs. Some PM2812 have 2 power modules, while some have 3. | ** Note that number of modules is not the same as number of outputs. Some PM2812 have 2 power modules, while some have 3. | ||
| Line 44: | Line 51: | ||
|3AT | |3AT | ||
"T" Typically means slow-blow fuse. Most of the time it's the first character in the code of a fuse, but here it appears to be used at the end of the code. | |||
Note that the following fuses have been found in actual power supplies, it is unclear if they came with these from factory: | |||
* On 3pcs of PM2812/4 originally configured for 110V operation 6.3A FERRAZ SU (medium time lag) fuses have been found | |||
* On 1pcs of PM2811, originally configured for 110V operation F2.5 (Fast Speed) fuses have been found. | |||
* On the bottom power module board there is a fuse, that is not documented in the service manual. The silk screen asks for T1.6A fuse (slow blow), but F1.6A fuses (fast blow) have been found in at least 2 places. (these fuses don't need to be adapted for 110V/220V operation, cause they see the same voltage/current independent of the supply voltage, so feel free to ignore this information) | |||
Latest revision as of 10:50, 23 February 2026
| PM281X: Switching between 110V and 220V | |
| Device | PM281X |
| Affects part(s) | |
| Needs equipment | Screwdriver |
| Type | General |
- 94V .. 130V (a.k.a. nominally 110V)
- 190V .. 255V (a.k.a. nominally 220V)
- Move the black jumper wire on the "Input Module" PCB, to the correct jumper. (this is marked via silkscreen on the PCB)
- (Note: the service manual forgets to mention this)
- See Figure 1 and Figure 2
- The blue (bl) wire of the cable loom DB1 must be connected to
- X4106 for 110V operation
- X4102 for 220V operation
- You can see the location of these pins on page 63 of the service manual [1]
- Also see Figure 3 and Figure 4 (note that the PCB from the factory has no text next to these pins, you can only identify them based on relative position)
- The fuses must be replaced. The required fuse current is different for each voltage configuration, and also depends on the number of modules the power supply is configured with
- Note that number of modules is not the same as number of outputs. Some PM2812 have 2 power modules, while some have 3.
- Please write on the back of the unit the current supported voltage.
Fuses
| Fuses 2x: | 1 Module PM2811 |
2 Modules PM2812/[0,1,2] |
3 Modules PM2812/[3,4]; PM2813 |
| 110V | 2.5AT | 4AT | 6.3AT |
| 220V | 1.25AT | 2.5AT | 3AT |
- On 3pcs of PM2812/4 originally configured for 110V operation 6.3A FERRAZ SU (medium time lag) fuses have been found
- On 1pcs of PM2811, originally configured for 110V operation F2.5 (Fast Speed) fuses have been found.
- On the bottom power module board there is a fuse, that is not documented in the service manual. The silk screen asks for T1.6A fuse (slow blow), but F1.6A fuses (fast blow) have been found in at least 2 places. (these fuses don't need to be adapted for 110V/220V operation, cause they see the same voltage/current independent of the supply voltage, so feel free to ignore this information)
